Dewpoint - different values. **solved**
Posted: Sat Oct 10, 2020 4:46 pm
by Lespius
Hi,
I notice that the dewpoint displayed on Raw Sensor Data (5.5°) is not the same displayed on MinMAX Data (6.0°):
The correct value for Temp=11.2° and Hum=68% is 5.5°
This is a problem because templates like Weather34 have the wrong value through their API (6.0°).
You can check Brian's answer:
viewtopic.php?f=56&t=14398&start=1140
Do you now what's wrong ?
Alain
Re: Dewpoint - different values
Posted: Sun Oct 11, 2020 11:43 pm
by admin
Images of your post is missing. No idea why values should differ and I am a bit skeptical that they really do.
Re: Dewpoint - different values
Posted: Tue Oct 13, 2020 8:58 pm
by Lespius
Ok, I will try to be more explicit.
On one hand you have this:
"Outdoor 5.9° 90% (
dew 4.4°, ...)" --> here dewpoint = 4.4°
http://www.meteo-cugy.ch/template/imgs/Capture32.JPG
On the other hand you have the Outdoor Dewpoint sensor --> here dewpoint = 4.9°
"Outdoor dewpoint 4.9°"
http://www.meteo-cugy.ch/template/imgs/Capture33.JPG
So we have two different values for dewpoint: 4.4° and 4.9°
I have checked, with a temperature of 5.9° and a humidity of 90%, the correct value of the dewpoint is 4.4°
So the value of the sensor is wrong. I must say that I have even seen it higher than temperature, which is physically impossible
Re: Dewpoint - different values
Posted: Sat Oct 17, 2020 4:35 pm
by admin
The yellow color of "outdoor dewpoint" indicates that this logical sensor is manuall redirected to a certain physical temp sensor. Therefore, I am not suprised that values do not match. Can you please show your mapping page?
Re: Dewpoint - different values
Posted: Wed Oct 21, 2020 9:15 pm
by Lespius
Re: Dewpoint - different values
Posted: Wed Oct 21, 2020 10:35 pm
by admin
Hm, strange. Could you please provide me a remote login? Please send URL and PW to "info(at)smartbedded.com"
Might be an old calibration infos somewhere in the settings file. But hard to say without having access.
Re: Dewpoint - different values
Posted: Fri Oct 23, 2020 9:37 pm
by Lespius
Done.
I will be away for a week. So don't be surprised if I don't react.
Re: Dewpoint - different values
Posted: Sat Oct 24, 2020 12:11 pm
by admin
Root cause is that there has been a calibration of -0.5°C defined for "th0temp".
On the "Raw Data" tab the dewpoint gets calculated from calibrated "th0temp" sensor, while on "Min/Max" tab the dewpoint is taken based on the original "th0temp". When calibration should also apply to "th0dew" it needs to be defined separately for this sensor. I admit that calibrating a calculated dew point value seperately is not best practice, but as Meteobridge does see all sensor data as unconnected data items it has be that way.
I added a calibration of -0.5°C to "th0dew" as well and data looks fine again.
Re: Dewpoint - different values. **solved**
Posted: Wed Nov 04, 2020 9:26 pm
by Lespius
Well done, thank you